Empower ethernet detector not switching from ready>running
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
3 posts Page 1 of 1
Empower 2154, SP/FR 4/D
2489 detector & 2414 hooked to a 2695. 2489 is ethernet, rest IEEE.
Last week, all was well. This week, using same method set, detector won't start collecting data on injection. It shows "Running" during monitoring and a baseline is displayed. When injections happen, it just stays on "Ready".
Any ideas?
Thanks,
DR
Did your inject start cable become disconnected?
Nope - I checked that.
I also created a new version of the instrument using just the 2695 & 2489 (there is also a seldom used RI detector available on that system), wrote a new instrument method for it, included it in the method set... still no go.
I'm starting to think maybe it's a hardware fault on the detector.
Thanks,
DR
3 posts Page 1 of 1