Flooded Quads, Noisy Baseline
Recently I performed an infusion with high amount of analyte ~10,000ng/mL at higher flow rate 30uL/min.
In my MRM method I have noticed that the background signal of mobile phase for my analyte, the one that was infused, is high ~1500-2000counts, whereby the I.S. which was not infused gives normal counts ~20-30. I have waited about 1 hour and have noticed counts dropped a bit for the analyte to 300-500, but is still much higher than normal (38) counts Any ideas, or advice as to and how to flush/purge the MS system? I think it is just a waiting game to let the N2 gas clear the ion transfer capillary and quadruples. I have also come to realize next time I will infuse only ~2,000ng/mL at a lower flow rate of 10uL/min.
Thanks.
Thanks.
Carryover can happen in a lot of places. Try replacing any tubing or fittings that you used for the infusion with new ones. You could also try flushing the system with LC grade isopropanol.
Users browsing this forum: DuckDuckGo [Bot] and 8 guests
