PDA wavelength Scan report on Empower
Posted: Wed Oct 24, 2012 2:30 pm
by StephenO
Hi All
I have a method which requires an identification of the active using the Water PDA Detector and the wavelenght scan for the sample has to match that of the Standard. I have done the analysis and can see the scans but How do I report these scans as in what do I select to report in the Reporting Method.
Thanks
Re: PDA wavelength Scan report on Empower
Posted: Thu Oct 25, 2012 9:04 am
by aceto_81
select them both, make a summary report and print the PDA spectra?
Re: PDA wavelength Scan report on Empower
Posted: Fri Apr 05, 2013 9:15 am
by fareedtayyeb
Hi there !!!
I am using Waters 996 with Empower software which is a good software.
Let me set the example for you.
Suppose i have a actual compound on wavelength is 230 with Mobile phase Meoh/h20 50:50.
Now we run the sample . The peak of interest appears with good resolution.
The question is that if somebody adds any impurity or anything other than my active compound in production process, how come i know that something is added in my sample .
Since i do not get any other response except my compound of interest with my mobile phase mixture at 230 nm.
Naturally the unknown impurity is not necesseraly will elute with my mobile phase combination and still i wish to know any other response within the spectral range of 200-400nm
Please guide me from this point
Thanks