Hi all,
I have some methods using ECD detector in amperometric mode and the signal considered is in the scale of nC. I thought in amperometry, the unit is in nA and in coulometry it is in nC. Am I wrong at some points ?
Thanks
Dear thohry
You are right amperometry usually meaures current (I) and coulometry measures charge (Q). But especially in Pulsed Amperometric Detection charge is quite often charge or integrated (over time) current is measured.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau