Units of amperometric and coulometric modes
Discussions about IC and related topics
2 posts Page 1 of 1
Hi all,
I have some methods using ECD detector in amperometric mode and the signal considered is in the scale of nC. I thought in amperometry, the unit is in nA and in coulometry it is in nC. Am I wrong at some points ?
Thanks
Dear thohry
You are right amperometry usually meaures current (I) and coulometry measures charge (Q). But especially in Pulsed Amperometric Detection charge is quite often charge or integrated (over time) current is measured.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
2 posts Page 1 of 1
In total there are 5 users online :: 0 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 5 guests