Page 1 of 1

32 karat Software Beckman Coulter HPLC. Failure of detector

Posted: Wed Feb 20, 2013 5:53 pm
by kaoutar
Hello,

I have a problem with the HPLC (32 karat software). I cannot calibrate the detector it always shows "calibration failed". It was working fine before, with minor problems but it's the first time I see this message.

Thank you for your help

Kaoutar

Re: 32 karat Software Beckman Coulter HPLC. Failure of detec

Posted: Wed Feb 20, 2013 8:23 pm
by skunked_once
It was working fine before, with minor problems
Do you mean the software was working before but with minor problems or that the detector was working before but with minor problems.

Which detector are you using?

I suspect that the problem is in the detector and not the software. Please post more information.

Also post this in the Liquid Chromatography section.

Re: 32 karat Software Beckman Coulter HPLC. Failure of detector

Posted: Wed Aug 16, 2017 3:45 pm
by OggIPN
Hello,

I have a problem with the HPLC (32 karat software). I cannot calibrate the detector it always shows "calibration failed". It was working fine before, with minor problems but it's the first time I see this message.

Thank you for your help

Kaoutar
Hi

Excuse me, there is any possibility you can share 32 karat software with us?

I am student of PhD in The National Polytechnic Institute in Mexico

I pay any cost of ship

Thanks so much
Best regards
Oscar Guzmán

Excuseme, there is any possibility you can share 32 karat software?