to use our electrochemical detector we need to perform a performance qualification to make sure everything is working fine.
We have some protocols for HPLC-PDA, but none for HPLC-ECD.
Anyone who knows what and how to test?
Kind regards
Ace
Hallo Ace
I would recommend to do dummy cell tests as described in the manual and further on run a calibration and repeatability on your compound of interest.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland