UPS - Battery Backups for 6890 GCs?
We have intermittent power interruptions at our facility for which we'd like to set up both of our 6890 GC systems on a UPS battery backup system. We have used these with good luck for our 1100 LC systems. Would this particular model support the power requirements of a 6890 GC?
http://www.amazon.com/APC-SMC1500-Smart ... B007ZT2KKM
The plan would be to set each 6890 system up on a separate UPS.
Most of the power interruptions we have only last for a few seconds. Thanks for any feedback from other users who have connected their GCs to a UPS system. I'd like to know which specific model(s) of UPS would be appropriate for a GC.
I have my 6890/5973 connected to an APC model SUA3000RM2U UPS and it provides ~10 minutes of backup which has eliminated previous shutdowns. It was a while ago so I don't remember how we calculated the size but maybe you can reference something from this model
Definitely need the 3000VA UPS for the GC oven- the 6890 is labelled for 2200VA max on the back next to the label. Watch the plug type when you order, they come with a 30A plug, which requires a new outlet in the lab.
Well worth the cost to protect your instrument, we have had quite a few power interruptions. Batteries last at least 5 years and easy to replace.
In total there are 82 users online :: 4 registered, 0 hidden and 78 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google [Bot], Semrush [Bot] and 78 guests
