Importance of Filtering MS Foreline Pump Exhaust ?
Posted: Sat Jul 13, 2013 3:43 pm
Hi folks,
I have read the few other posts that I found addressing this, but would value any opinions on my specific circumstance (methods run, freq. of runs).
I was gradually promoted in my lab about 5-6 months ago to operating our GC (5890), GC/MS (5890/5971 & 6890/5973), and HPLC. I have been really enjoying it, learning a great deal, and excited versatility of these instruments in terms of career opportunities (I graduated w/ my B.S. one year ago). Aside from the HPLC, the instruments seem to have been somewhat neglected the last few years. I have been using this forum frequently in combination with a few books and operating manuals to assist in troubleshooting and have been successful in correcting simple issues that were causing serious inefficiencies - thank you all particularly Peter Apps and Don Hilton!
The foreline pump (old Edwards model?) on our 5971 is not vented to a fume hood and does not have an oil mist filter. We run methods 525 and 548 once a month as well as 624 and 625 probably twice a month. The oil has not been changed in probably 3 years or more. I requested that the company install an oil mist filter, but was denied as to them it is a waste of money unless you are running dioxins for a superfund site. After writing a lengthy email explaining my reasoning, citing sources and the specific chemicals of concern, I was told ......
"It is my opinion that the amount of chemicals released into the air from the am is "di minimis". It is only one 1ul injection per hour and anything that does make it through the ms spec is immediately diluted in the lab air system. In terms of risk you probably inhale 1000x more benzene and other chemicals ever time you fill your car up with gas. But if it makes you happy we can get a filter for the 71"
And as far as my efforts to develop and follow a routine maintenance schedule (changing septa and liners, clipping front end of column, updating manual tune), I was told we follow an "if its not broken lets not fix it approach".
Am I being unreasonable and overly cautious?