vitamin B on HILIC
2 posts Page 1 of 1
I'm trying to develop an analysis method to determine vitamine B1, B2... with HPLC-MS
With a normal C18 column i had a lot of tailing so i decided to switch to hilic as some of the analytes were rather polar.
I'm using an xbridge HILIC 3.5 µm column (2.1 x 100 mm) and the solvents are: A: 0.1 % FA in ACN
B: 0.1 % FA + 10 mM NH4OAc in H2O
I'm using a linear gradient that goes from 5% B to 50 % B in 10 min.
With my method i have a rather good separation of all my analytes but I do have a lot of peak broadening for my vitamine B6 (Pyridoxal)
The peaks before B6 and after B6 are OK.
Can anyone help me to get rid of this peak broadening problem?
Best regards
here is an alternative approach for Vitamins:
http://www.sielc.com/Compound-Pyridoxine.html
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
2 posts Page 1 of 1
In total there are 9 users online :: 2 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Google [Bot] and 7 guests