AutosystemXL Error
Greetings,
Has anyone ever encountered an error on their Autosystem XL that reads as follows:
"No DTF3 Flow Module at Connection 5.'
Thank you for your assistance.
Cyprian
DTF is the detector flow. It is not recognizing the PPC module (i believe position 5 may be for the air?) on the PPC board. Make sure the module cable connector is seated well on the PPC board. You could have a bad PPC module, PPC board or even a Main board.
It was the main board along with the PPC module. Thanks for your comment!
Hi
I have the same issue on my Perkin elmer GC 'Warning CAR 1 pressure less than 50 psi'
Please advice.
Hi
Please advice.
open your cylinder valve and set it to 100 PSI.
then, check for leaks, and you'll fix it.