Septum purge shutdown
GC/MS 7890 has been shutting down due to inlet pressure. It seems unable to purge thus will not move onto next vial.
I have changed vent copper line and filter and it ran 20 vials then the same thing (liner/septa, septa nut all changed as well).
has anyone had a problem with the analogue valve malfunctioning so that it is physically unable to open valve? Any suggestions greatly appreciated.
Is the pressure too high or too low ? Have you checked for leaks ? - using a leak seeker not the onboard diagnostics.
It is too high. Hovering at 32-33psi. Previously started run at 17.5.
No leaks found
A stuck back pressure regulator on the split is a distinct possibility. Check for flow out of the split vent. What do the pressure and the split flow do if the inlet is in splitless mode ?.
In total there are 34 users online :: 0 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 34 guests
