too much noise
sth wrong with my varian detrctor 320,the noise was too high about 100mv,days ago the detector was blocked and I fixed it by flow some
hac and pure water through the flow cell ,but after that time ,I find that the noise was too high ,who can tell me what can I do?
...clean the flow cell, it seems that there are some waste particles bouncing around....
regards Chris
but I do not know how to clean the flow cell ,is there any procedure?can I ultrasonic the flow cell
First remove flow cell
then have a look through the flow path(s)while shaking it slightly
If you can see some dirt or the lenses are not clear, flush it with organic solvent, if this doesn´t help, you can use 1n HNO3
Then flush it with eluent aggain.
regads and good luck
Chris
Flowcell cleaning procedure:
Flush with methanol followed by water to remove all of the methanol.
Flush with up to 30% nitric acid. Leave the acid in the flowcell for 5-10 minutes then continue flushing.
Flush with water until ALL of the acid is removed then flush again with methanol.
If this doesn't work, you will have to find the detector manual and take apart the flowcell to clean the windows. Be aware that you may need to buy new flowcell windows and gaskets.
Users browsing this forum: John Guajardo and 14 guests
