difference between split and splitless liner
Posted: Fri Sep 27, 2013 10:01 pm
why we need to make two different types of liners?
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://chromforum.org/
You are doing two different kinds of things in the inlet.
In one you are trying to get everything you injected to transfer to the column in a narrow band with the time required to transfer material to the linlet taking long enough to transfer that volume of gas from the inlet at the flow rate of the column. Mixing in the inlet makes the transfer longer. (splitless)
With the other you are sampling from a fast moving stream of gas as most of it heads for the split vent. You need adequate mixing so that you get a good sample from that gas. Time is not as much of an issue because with the gas flow going out the split vent, the inlet is quickly cleared of sample. (split).