is there any problem with injecting DMSO LCMS system?
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://chromforum.org/
If you inject a small amount of a reasonably dilute solution of dmso, such as happens if you have a standard chemical solution in DMSO at high concentration, that you have used to prepare standard curve dilutions in a more friendly solvent, then No, there is not a problem.
If you inject larger amounts, yes, it will linger for ages and ionise strongly, which may be annoying, especially if (1) you are looking to collect good full-spectral data; (2) you are using an instrument where sensitivity is limited by total number of ions, e.g. many ion-traps; (3) you happen to be analysing something with the same mass as DMSO or its cluster ions.