NPD with H2 Carrier
Howdy!
I'm curious if anyone has tried using hydrogen as a carrier gas with a nitrogen phosphorus detector. I recently switched my 6890 from helium to hydrogen due to cost and availability. I haven't been able to mess with it much yet, but I thought I should see if anyone has some experience I could learn from.
Thanks!
It should work fine as long as you keep your total H2 flow to spec which is 3ml/min (5 max) so most of that is coming from the column carrier flow.
Good point. My total flow was 5.5 mL/min and I was having trouble igniting. I dropped the total down to 3 mL/min and it's igniting like a champ.
In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Users browsing this forum: Semrush [Bot] and 24 guests
