QuEChERS - high background after cleanup in GC-MS
Posted: Thu Jan 16, 2014 10:59 am
After blank (only water) QuEChERS extraction & clean up I get a very high background in GC-MS scan Mode.
So it seems that I extract a lot of compounds out of the used plastic ware.
Is this normal?
Any recommendations?
I use the SampliQ QuEChERS kit from agilent.
Extraction Kits: contains 4 g MgSO4; 1 g NaCl; 1 g NaCitrate; 0.5 g disodium citrate sesquihydrate
Cleanup Kit 1: contains 150 mg PSA, 900 mg MgSO4
Cleanup Kit 2: contains 150 mg PSA, 150 mg C18E, 900 mg MgSO4
Cleanup Kit 2 gives less background.
In both kits "Sarstedt" Tubes are used.
GC: Shimadzu 2010plus
MS: Shimadzu QP2010 Ultra
LVI using ATAS Optic-4
Injection 50µL
Split 1:100
Split Ratio Program
0.01 0.0
2.00 50.0
Puge Flow 3.0 ml/min
Carrier Gas 1.2 mL/min He
Oven
60°C to 130°C at 10°C/min
to 230°C at 4°C/min
to 300°C at 10°C/min
hold for 10 min
MSD
Interface: 300°C
Ione Source. 250°C
Solvent Cut 5 min
Voltage 0.2 kV (relative)
SCAN:
45 – 500 m/z (0.2 sec)
Autosampler
Air Volume [µL]: 0
Pre Clean with Solvent 1: 0
Pre Clean with Solvent 2: 2
Pre Clean with Sample: 1
Filling Volume [µL]: 50
Filling Speed [µL/s]: 10
Filling Strokes: 5
Pullup Delay [ms]: 5000
Injection Speed [µL/s]: 75
Pre Injection Delay [ms]: 500
Post Injection Delay [ms]: 500
Post Clean with Solvent 1: 0
Post Clean with Solvent 2: 5
Optic-4
Hot Splitless
Equil.-time: 5 sec
End time: 960 sec
Initial T: 60°C
Delay: 5 sec
Ramp: 15°C/sec
Hold T1: 280°C
Hold time: 939 sec
Vent Mode: fixed time
vent Time: 50 sec
Is this normal?
Injection 50µL
Split 1:100
Split Ratio Program
0.01 0.0
2.00 50.0
Oven
to 230°C at 4°C/min
hold for 10 min
MSD
Interface: 300°C
Ione Source. 250°C
Solvent Cut 5 min
SCAN:
Autosampler
Air Volume [µL]: 0
Filling Strokes: 5
Optic-4
Hot Splitless
Equil.-time: 5 sec
End time: 960 sec
Initial T: 60°C
Delay: 5 sec
Ramp: 15°C/sec
Hold T1: 280°C
Hold time: 939 sec
vent Time: 50 sec