Dionex ICS-2000 backpressure restrictor coils
I'm running on a Dionex ICS-2000. I recently needed to change out the guard column bc I was losing resolution. Once I did this the pressure was not as high as it should be (2000-2300psi). I know you can add backpressure restrictor coils to help with this but the ones that I currently have are not bringing pressure as high as I need. I've tried red and yellow gold coils. If anyone has any hints or ideas it would be great! thx...
Hi contestlabs,
Some questions:
Some questions:
- What was your pressure before and now after the guard change?
- What column set are you using?
- What is your flow rate?
- How many coils of red and/or yellow tubing do you have installed between the EG cartridge and the injection valve?
- You can link them with a black coupler if needed to achieve higher pressures.
Best regards,
John Guajardo
Product Manager
Thermo Fisher Scientific
John Guajardo
Product Manager
I'm using IonPac AG18 guard 4x50mm and Ion Pac AS18 column, @ 1.0mL/min using ECG KOH @ 32mM and suppressor at 80mA 4x250mm.
The pressure before I changed the guard was ~2100-2200 psi; however I was losing resolution on column and QCs failing; on inspection of the guard and column inlets i noticed the guard had lost headspace (column looked fine), then I replaced the guard.
Once I replaced the guard the pressure was around 1400psi with one red coil; I tried a gold coil but it was too high, trimmed gold coil a bit to see if it lowered, still not low enough...Currently I have 2 red coils together with a black coupler btwn the injection value and eluent out and the pressure is around 1700psi.
Perhaps I need to give the system more time?
Thx
Thx
There are two types of red coils and two types of yellow coils in the shipkit:
P/N 053764 is yellow tubing that gives 500 psi @ 0.5 mL/min
P/N 053765 is yellow tubing that gives 1000 psi @ 0.5 mL/min
P/N 053762 is red tubing that gives 500 psi @ 2.0 mL/min
P/N 053763 is red tubing that gives 1000 psi @ 2.0 mL/min
If you don't have the p/n, there should still be a label clipped to the tubing advising of the pressure @ x flow rate.
Based on your current pressure of 1400 psi, coupling the two red tubings together (053762 and 053763) should get you an additional ~750 psi @ 1 mL/min and bring your overall pressure to ~2150 psi.
Best regards,
John Guajardo
Product Manager
John Guajardo
Product Manager
Thanks John for your help!
I think I need to order another type of red coil and that should get the pressure where it needs to be...
Are you in Massachusetts? Perhaps we can collaborate. Also, I have loads of spare tubing and parts in my lab. I'd be happy to share what I can.