Empower2 Processing Error
8 posts Page 1 of 1
Has anyone ever encountered an error where Empower processes the results individually when Batch Processing had been specifically selected?
Batch processing is supposed to generate individual result files in addition to a result set file.
DR
The real issue is that a result set was never created even though batch processing was performed.
Check the projects table space quota and the message center...
DR
This happened right after more space was added to the table. Is that what caused this?
Usually, it won't batch at all if you're short on table space.
Since you added table space, batch it again. If you get a results set this time, it was most likely lack of table space causing the problem.
DR
As usual: what's in the message center?
after adding more table space, did you close and restart the project before processing?
do you process the sample set or just seleceted injections? In the latter case, no result set will be produced, just single results