We are trying to qualify a Waters Alliance instrument with a 2998 PDA detector. We are wanting to do a single PDA scan of a NIST holmium oxide standard. Can anyone provide the steps we need to acquire the sample through Empower 2? I am having a brain fart on how to do it. Thanks!
Make a 0.0 uL injection of a sample from within Empower using your desired sample set for your desired length of time.
You will probably want to use a PDA scan of the "sample" for about a minute or so.
2 posts Page 1 of 1