I just got a very old HPLC (HP 1050 series), and I'm trying to connect it to chemstation.
There is no HPIB (or GPIB) port on the back panels of HPLC. The autosampler, pump and detector are connected together with remote cables. And there is a analog cable that outputs signal from detector to an integrator.
I might need to install HPIB (or GPIB) output board on my HPLC to achieve the connection. But how many output boards do I need? Only one board for the detector, or three for detector, autosampler and pump?
Also, where could I buy these kind of boards? They are really old! I tried ebay, but could find it.
Thanks!!! I really appreciate your replies!!!
M
I am pretty sure we have a set of HPIB boards for a 1050, give me a call and we can figure out what you need.
Alex at 303-730-4136 ext. 905
Best Regards,
Alex Hummell
Analytical Instrument Management
ph. 303.730.4136 ext. 902
cell. 720.203.5224
http://www.AIManalytical.com