delta p for merck HPLC column
hi all
the CoA for merck lichrocart says something like this
test results for toulene
k=1.9
N/m=82789
As=1.0
delta p=64 bar
what does this delta p stand for ...?
hoping to get some information ....
with best wishes
HPLC System: HPG system with two(2) Knauer S1000 pumps- Dyamic Mixing Chamber-Knauer s3900 auto sampler-S2300 RI detector-S2500 UV detector- Raytest radioactivity detector
tarzan
usually back pressure observed with CoA mobile phase
Vlad Orlovsky
HELIX Chromatography
www.helixchrom.com
hi thanks for the quick reply
the CoA says 60 bar for 75:25 ACN water ..
but I get 90 bar ...??
what should i do ..?
bets wishes ...
tarzan
See how much back pressure you have without column. May be the difference comes from your system
Vlad Orlovsky
www.helixchrom.com
An external manometer after the column will give you the exact pressure. Maybe the built in electronic pressure damper is not set properly!
Can be set by the service technician. Good luck.
