PH of dileunt and M.PH. for ciprofloxacin
3 posts Page 1 of 1
hi everybody ,
nowadays i am using a pharmacopoeial hplc method for analysis of ciprofloxacin HCl ,
mobile phase ammonium phosphate buffer (PH 3 ) : ACT 87:13
but sample dilutions are made by diluent ammonium phosphate buffer ( PH 2) : ACT 87:13
i can understand the reason for mobile phase ph ( 3 ) and its relation with ciprofloxacin dissociation constants knowing that ciprofloxacin molecule has -cooh gp and nitrogen on piperazinyl ring ( pKa = 6.09 (carboxylic acid group); pKa = 8.74 (nitrogen on piperazinyl ring)
what i cant understand is why the sample dilutions cant be in mobile phase ( ph 3) and what is the reason for the diluent ph ( 2 ) ,
we didnt use ph ( 2 ) as mobile phase as a protection of column packing form this PH ( hydrolysis of silanol gps, right ?
i hope if anyone can help me with any information
thanks to all
Don't know what Pharmacopea you use, but in the Ph. Eur. the diluent is equal to the mobile phase (pH 3.0)
Ace
thanks alot for replying , it is according to USP 35