Gradient separation: Theoretical gaps
Posted: Tue Aug 05, 2014 5:44 pm
I have a gradient method:
20% MeOH, pH 2.5 => 80% MeOH, pH 2.5 in 6 minutes
Flow: 1.2 mL/min
Column dimensions: 150 x 4.6 mm (column volume: 1.7 mL)
Retention times: 3.5 minutes for compund A and 5.3 minutes for compound B
(after this program: in 2 minutes back to 20% MeOH pH 2.5 and reequilibrate for 2 minutes; total time = 10 min)
According to JW Dolan's equation (http://www.chromatographyonline.com/lcg ... =&pageID=2), I can calculate k* assuming a value of 5 for S:
k* = (6 * 1.2)/(1.15*1.7*0.6*5) = 1.23
In isocratic separations, k' should ideally be between 2 and 10. Is the same also true for k* in gradient separations and thus, should I modify a part of my variables (e.g. using a 2.1 mm column or start at 40%)?
Additional questions:
- why do retention times are not part of the k* equation?
- if I apply the approach of S = 5: does k* characterizes my gradient system in general?
- if I want to calculate S for both of my compounds out of a log k versus % MeOH diagram (linear): how to do this?
for example: 37.5% MeOH results in k = 20, log k = 1.3 / 48.5% MeOH results in k = 2, log k = 0.3
=> delta of log k = 1, delta of % MeOH = 11% = 0.11
=> S = 1/0.11 = 9 ???
Thank you very much for your help.
Regards
Florian
Flow: 1.2 mL/min
Regards
Florian