ICS2000 columns cleanup

Posted: Wed Oct 08, 2014 4:46 am
by Aster
I performed a columns cleanup (AS-11-HC guard and separator) followed the instruction in the manual, to solve a problem of reduced retention time and floride tailing. The cleaning solution I used is 200mM in 80% Actonitrile. It didn't help. Any advices?

I also found that it took so long (>1day) to bring the background to be less than 1uS after the cleanup. Is this normal?

Thanks,
-Aster

Here are chromatograms showing my problems with retention time.
https://www.dropbox.com/s/seimgjp0j7pn4 ... e.xls?dl=0

Thanks!

Re: ICS2000 columns cleanup

Posted: Fri Oct 17, 2014 2:07 pm
by aemenari
Your column is probably old. Lowering the Eluent concentration and flow rate may improve peak separation.

Re: ICS2000 columns cleanup

Posted: Mon Oct 20, 2014 3:58 pm
by Aster
Thanks for comments! Our columns are quite new, used for only about 400 analysis ( but some of the samples have high salinity). A lower eluent con helps a little while reduced flow rate doesn't.