internal standard
Hi
1-I use capric acid (ester)as internal standard in analysis of methyl oleate.
but when the area of analyte increase, area of ISt deacrease or conversely.
I haven't any explanation for it.
2-how do I calculate the response factor?(is it correct: area/amount of analyte)in quantitation analysis,should use the response factor?
please explain for me
thanks in advance
Your results do not appear authentic. I would suggest that you explain your procedure in more detail. Perhaps then a reasonable explanation for your unusual results can be postulated.
Hi
internal standard:methyl decanoate
analyte:methyl oleate
column:PEG 20M capillary column,30m,0.22mm id,
carrier gas and make up:He 1.53mm/min
hydrogen and air:30,300 mm/min
det temp(FID):250
inj temp:240
column temp:160,5C/min,230
split ratio:1:50
conc istd:50ppm
thanks
You do normal external standard calculations EXCEPT that you use the ratio of your analyte/internal std peaks instead of your analyte area everywhere in the calculations.
DR
thanks alot
nina
Users browsing this forum: Bing [Bot] and 8 guests