Peak Purity
Is it possible for a 100% pure compound to produce a peak purity <100%? What could cause this? Could operating at the pKa of one of the substituent groups cause lower than expected peak purity due to the compound existing is 2 different ionized states? Is PP a useful parameter?
thx
AFAI see ideal peal purity at Agilent 1260 is always ≥ 99.9 % but I can't remember 100 %…
Best regards,
Dmitriy A. Perlow
Hi Mike H.,
Mike H. asked:
Yes, it is certainly possible for a "100% pure" compound that, when chromatographed, affords data that indicates that its chromatographic peak is not spectrally homogeneous. I'll add that it is also possible for a "<100%" pure compound that, when chromatographed, affords data that indicates that its chromatographic peak Is spectrally homogeneous.
By spectrally homogeneous, I mean that, within the noise of the HPLC/photodiode array measurement system, that all acquired spectra over the selected range of spectral acquisition for the analyte concentration and at the given separation conditions at that time are indistinguishable from each other.
What could cause the "100% pure" compound to not afford data for a given separation that indicates spectral homogeneity? Mike, you note one cause above, incorrect software settings could also lead to a similar result (particularly with gradient separations and the peak tails vs. the peak apex). Each software package has its own idiosyncrasies, it seems to me.
Is Peak Homogeneity a useful parameter? I'd say yes if it is properly applied, executed and its limitations understood.
Mark Stahl wrote this (Agilent .pdf): 5988-8647EN
Arvind Mittal did a nice YouTube video:
https://www.youtube.com/watch?v=E-Whxm398BE
CHROMacademy wrote this, you'll have to sign up for a Lite membership to view the entire document, sign up is Free.
http://www.chromacademy.com/chromatogra ... ctors.html
MattM
Thx. this is extremely helpful.
:P
MattM
In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)