Cremphor HPLC troubleshooting
Dear all,
I am developing an eye drops solution that contains Cremophor RH40 and Cremophor EL in percentages form 0.25% - 5%.
When analyzing the product in normal-phase HPLC with silica columns, i have noticed that the life time of columns is decreasing dramatically, thus making it difficult to have trustworthy results.
What i see is that the retention time of the main peak is moving towards smaller retention times and that the baseline becomes smoother.
I wonder and would like to ask the following:
a) Is there a way of protecting the silica column (eg. Use of precolumn or a special cleaning treatment that removes cremophor or something else)?
b) Is there a special treatment for removing cremophor from the final eye drops formulation, before preparing the impurities sample for HPLC analysis?
Thank you in advance!
If the problem is really accumulation of your surfactant on the stationary phase then yes, a guard cartridge should help. You will have to do a loading study on the cartridge to establish how often it should be replaced.
Solid phase extraction (SPE) is a possibility. My first guess would be to use a reversed-phase cartridge, but there are a wide range of chemistries available and it's easy to try different ones.
Users browsing this forum: Semrush [Bot] and 6 guests