Cold Trap O-ring Change Markes Thermal Desorber TD-100
Hey,
I am planning on changing the cold trap oring for the first time on my markes TD-100. Wondering if you of you might have done this before. For those who are familiar, this is not the oring connected the removable solenoid assembly. This is on the opposing side. i would says it seals the top of the trap, its closest to the weak sorbent bed. It appears the entire peltier must be removed or at least pulled back to access the oring. Just wondering if anyone can confirm this is true or any pointers to successful completion. This procedure seems to not be mentioned in the user manual.
thanks all!
Hello
It is tricky operation but it is all about practice.
All you need is:
- o-ring extraction tool (with sharp tip)
- o-ring insertion tool (with small ball at the end)
- new o-ring
- filter disks
And a lot of patience :)
Good luck
Regards
Tomasz Kubowicz
interesting and unfortunate at the same time :lol:. thanks for taking the time respond, i do appreciate it. I am feeling patient today gonna give it a stab.
edit: apparently i have two forum usernames...