6850 GC Flame does not stay lit
Posted: Tue Jan 20, 2015 5:15 pm
My department has an HP 6850 GC with FID that is used only once or twice a year for an undergraduate teaching lab and is shut off completely the rest of the time. It was last used in January of 2014. I turned everything on two weeks ago to get ready for this year's class: checked for adequate gas supply, lit the FID flame, let the instrument run for ten minutes before shutting off, and thought all was well. No parameters had been changed since last year; no one has touched the instrument since last January.
On the first day of class last week, as I was explaining the instrument operation to the students, the FID flame went out for no apparent reason after about 10 minutes. I could not get it to re-light. I tried changing the proportion of gases, with no effect. According to the ChemStation readings, gas flows for H2, air, and He came up to their set points (H2 30, air 300, make-up He 30 mL/min) with no problem, as usual. I heard the "pop" of the igniter in the FID, and I could see the element glowing hot. Just no flame on.
I thought to check the column for breaks or poor connections to the inlet/detector, but this turned up nothing.
A few days later, when I had time to go back to this, I was able to light the flame on the first try (using the original settings). The flame stayed on for exactly 14 minutes and then went out, again for no apparent reason. The instrument automatically tried to relight and succeeded. The flame stayed on for about 1 minute and then went out again. Again the instrument did its automatic relight procedure, but this time nothing. It tried maybe four cycles (each time I heard the "pop" sound and saw the igniter glowing) with no effect. I tried turning the flame off and then back on "manually" (by clicking the command boxes in the software) but that didn't work either.
We were using an H2 generator to provide hydrogen, so I thought perhaps there was a problem with that. I disconnected the generator and substituted compressed H2 from a cylinder (air and helium have always been from cylinders). Again, no trouble lighting the flame on the first try. Stayed on for 14 minutes, then off for no reason. GC went into its reignite routine and was able to relight and stay lit for about a minute, then not after that.
I don't know why the flame goes out after 14 minutes. If I can light it in the first place, why doesn't it stay lit? I'm not getting any kind of error message or indication of any problem at all until the GC gives up trying to reignite (then, of course, it automatically goes into a safety shutdown). I would be grateful for advice and suggestions.
On the first day of class last week, as I was explaining the instrument operation to the students, the FID flame went out for no apparent reason after about 10 minutes. I could not get it to re-light. I tried changing the proportion of gases, with no effect. According to the ChemStation readings, gas flows for H2, air, and He came up to their set points (H2 30, air 300, make-up He 30 mL/min) with no problem, as usual. I heard the "pop" of the igniter in the FID, and I could see the element glowing hot. Just no flame on.
I thought to check the column for breaks or poor connections to the inlet/detector, but this turned up nothing.
A few days later, when I had time to go back to this, I was able to light the flame on the first try (using the original settings). The flame stayed on for exactly 14 minutes and then went out, again for no apparent reason. The instrument automatically tried to relight and succeeded. The flame stayed on for about 1 minute and then went out again. Again the instrument did its automatic relight procedure, but this time nothing. It tried maybe four cycles (each time I heard the "pop" sound and saw the igniter glowing) with no effect. I tried turning the flame off and then back on "manually" (by clicking the command boxes in the software) but that didn't work either.
We were using an H2 generator to provide hydrogen, so I thought perhaps there was a problem with that. I disconnected the generator and substituted compressed H2 from a cylinder (air and helium have always been from cylinders). Again, no trouble lighting the flame on the first try. Stayed on for 14 minutes, then off for no reason. GC went into its reignite routine and was able to relight and stay lit for about a minute, then not after that.
I don't know why the flame goes out after 14 minutes. If I can light it in the first place, why doesn't it stay lit? I'm not getting any kind of error message or indication of any problem at all until the GC gives up trying to reignite (then, of course, it automatically goes into a safety shutdown). I would be grateful for advice and suggestions.