Chemstation Custom Reports adding %RSD and correlation coef
Hello!
Anyone out there that can can provide some insight using Chemstation Custom Reports to list the correlation coefficient for the individual compounds calibration curves.
I found the feature to add the avg RSD for each compound, however I do not see the selection to display the calibration coefficient if that was how the initial curve was constructed. For some reason, the avg RSD will still be displayed, even though the calibration for a particular compound is based on coefficient.
The version of Chemstation is 1701BA running under Windows 2000; HP 5890/5971 system analyzing EPA 8260B.
Any help to point me in the right direction is greatly appreciated.
THANKS!
Better Living Through Chemistry...
Under the Initial calibration menu there is a tab called 'List Compounds' which lists the copounds with either RSD or R^2 coefficient, but i think you need a macro installed to run it.
Check in the MSmacros folder and look for list.mac
I'm not sure if it is a macro that comes with chemstation or if it is one that we have written ourselves.
----suffers separation anxiety----
That's a feature of the older software. To get around it set up a custom tool. Where you enter text, type this.
macro _macropath$+"enhviewm.mac",go
Then when you are in DA hit the icon for the custom tool you set up. A box will open choose calibration report and you will get a R2 report.
Users browsing this forum: Bing [Bot] and 1 guest
