CompiPal Autosampler Problem with Saturn GC-Ion Trap
Posted: Tue Jan 27, 2015 10:48 pm
by tima
Hello Bright Ladies and Gentlemen,
We have a CompiPal that has broken it's SPME syringe after the adsorption step has completed. The AS raises itself out of the vial, it doesn't fully retract up. Then as the AS moves to the injection phase of the process and bends the needle and breaks the SPME fiber as it attempts to move to the injection port.
These are expensive things. So we are struggling to find where in the documentation can you calibrate, check or troubleshoot this problem.
Any suggestions would be greatly appreciated.
Timothy
Re: CompiPal Autosampler Problem with Saturn GC-Ion Trap
Posted: Wed Jan 28, 2015 8:25 am
by Forse
Worn elastic bands in the combipal.
Posted: Wed Jan 28, 2015 12:01 pm
by Peter Apps
Is (only) one of several possible causes.
First you need to check the alignment at every position that the syringe has to go to.
Check that the shaft of the SPME holder is not sticking (if the fiber is not retracting all the way), and that the syringe is properly mounted on the carrier.
Some types of vial septum are difficult for the needle to penetrate, butyl rubber is the worse.
Check that you are not overtightening the septa of the headspace vials or the inlet.
This is a perennial problem and there is lots in the archives - try a search.
Posted: Wed Jan 28, 2015 12:26 pm
by BMU_VMW
First check I would made are the elastic bands to.
Posted: Wed Jan 28, 2015 6:08 pm
by tima
Thanks for all the responses .... we ordered replacement injector bands.
Timothy
Posted: Thu Jan 29, 2015 8:17 am
by Forse
Because the old ones were broken?
Posted: Fri Jan 30, 2015 10:22 am
by BMU_VMW
if you need a quick fix just te keep you up and running until the replacements arrive you can tie a new knot in the band, just to make them a little bit shorter.
Posted: Fri Jan 30, 2015 11:24 pm
by tima
Not broken but stretched....iImothy