Hi
I'm setting up a used Waters GCT GC-TOF.
I need proper hardware interface as well as software and documents for that. Please let me know if you've anything in hand.
thanks and regards
bahram111@yahoo.com
GCT will need both the TDC and optical loop card installed in a EPC with bootable VxWorks connecting to host running Masslynx.
Hi
I'll be glad if you give me idea about this connection and it's requirements. I've GCT classic +Agilent 6890 GC, purchased a rough pump and have pirani gauge for rough.
I think I've to find a TDC and Optical loop cards too. Please let me know their details and needed software or any important part.
any document and resource is highly appreciated.
thanks and regards
bahram
bahram111@yahoo.com
Talk to Ray at mass-spec detectives; he may have the cards you need. Hook up is fairly simple and the MCP condition will be important. Note position of the Pirani on the GCT series rotary pump and the switching isolation valve associated with the rotary pump Tee-it wires back to the GCT itself to protect the TOF when changing sources.
Thanks for your recommendation. I've wrote him about that and hope to get reply.