How to check the TCD detector broken or not?
Posted: Sun Apr 12, 2015 11:55 pm
by LUXIAOXIAO
I tried measure the resistance and I can hear the clicking sound. Is there any other way to check?
Thanks!
Re: How to check the TCD detector broken or not?
Posted: Mon Apr 13, 2015 1:48 am
by tangaloomaflyer
Do you get a signal when you run a standard?
Re: How to check the TCD detector broken or not?
Posted: Mon Apr 13, 2015 2:09 am
by LUXIAOXIAO
Hi,
Thanks for your response and help.
When I turn on the TCD detector, the signal turn from straight line to noise. When I increase the mass flow rate from 18 to 25, the noise position changed from about 0.1 to 0.2. I don't have standard gas, I just use 2 percent oxygen in Ar. There is no peak. I used 2 percent C3H6 in Ar, still no peak.
Another question, will the reference and make up gas flow rate affect a lot? I just wanna see a peak.
Thanks!
Re: How to check the TCD detector broken or not?
Posted: Mon Apr 13, 2015 4:39 am
by tangaloomaflyer
How are you introducing your sample gas to the GC?
Re: How to check the TCD detector broken or not?
Posted: Mon Apr 13, 2015 5:23 am
by LUXIAOXIAO
Six way valve, this valve works since I used it for FID detector and there is peak.thanks!