I'm trying to set up Empower 2 to do the following with standards and check standards.
1. Calculate the mean and RSD of the first five standards.
2. Calculate a check standard relative to the initial standards.
3. Calculate each bracketing standard relative to the initial standards.
4. Calculate each sample based on the mean of all standards, including the bracketing standards.
Is this possible within Empower 2, without processing everything twice? It seems, on trying to test it out yesterday, that we could get the bracketing standards to act as controls (point 3) or as standards (point 4) but not both.
If it can be done, how?
Yes, this is possible.
Setting this up depends upon your use of custom fields and sample labels.
I have set this up for many companies in the past and would be willing to help out in this area as I am working as an Empower consultant.