caliberation curve slop keep on decrease
8 posts Page 1 of 1
Hi, I use API 2000 ms/ms to do a drug qutitative analysis. but the caliberation curve slop keep on decreasing although I do use a internal standard. :cry: Anyone can help? which part , LC or ms is resposible for it? Thanks.
emily lee
i would suspect that it is the mass spec. Possibly need to clean the source?
Sailor
Thanks. I have clearned the MS, problem still there. Any other suggestions?
emily lee
Do you mean that the curve is nonlinear, with decreasing slope at higher concentrations? Or do you mean that on multiple analyses of the calibration curve, the slope of the line keeps decreasing?
MG, thanks for reply. It is the later one, decreasing when comparing multiple caliberation curve, the later aquired one always has a lower slop that the earlier aquied one.
emily lee
What is your matrix? Also, what about QC standards interspersed with your samples: are they generally lower than the theoretical value?
Hi Meillid,
Is it an old or new MS method? I need more information about the condition MS and LC method. Could be possible to have a MS method robust problem.
Alejandro
Applications Chemist in Mass Spectrometric
hi, Thanks all for the replies. I found out the problem. the outlet line of the injection port is gradully cloging when the pump was finally shut down by high pressure, after replace the line, caliberation curve come back to normal. Thanks again for all the discussions. I really enjoy here.
emily lee
8 posts Page 1 of 1
