Peak purity vendor qualification (either LC-MS or LC-DAD)
On reviewing our LC-MS vendor qualification, QA asked me how do we prove peak purity testing through actual qualification testing. I told them that I wasn't aware of any vendor that would be able to do this vigorously. So the obvious question came up - how would we know the system is fit for its intended purpose?
Does anyone know if any vendor would qualify peak purity testing either for a LC-MS or a LC-DAD? I guess not, but I could be surprised ;-)
Hello
Part of IQ/OQ for software (eg.Chemstation, MassHunter) is test called "Spectral Library/Evaluation test". This test verifies that you can create a library, add a compound to it, and then identify the compound and verify the
identification.
So if you have IQ/OQ documentation for your software, you can prove it is qualified.
I don't think you will have specific procedure for "peak purity". But above tests can prove that spectral module is working and "peak purity" is part of it.
Regards
Tomasz Kubowicz
Users browsing this forum: Google [Bot], Semrush [Bot] and 6 guests
