High Percent RSD for vinyl chloride
Trying to resolve problem with vinyl chloride failing % RSD. Response factors decreasing with increased concentration. Any ideas?
Hello
You didn't write anything about method:
-parameters for GC
-sampling technique - Purge and trap?
More details you give, quicker answer you'll get
Tomasz Kubowicz
-parameters for GC
Tomasz Kubowicz
What is your main problem, RSD or response factors ?Trying to resolve problem with vinyl chloride failing % RSD. Response factors decreasing with increased concentration. Any ideas?
Relevant details neccessary if you want anybody to devote time for your problem.
The main problem appears to be the decreasing response factors which is causing the %RSD failure. We are getting linear curves for all gases, except for vinyl chloride. The responses for vinyl chloride are linear for the first 3 calibration points and then start declining from the midpoint to the highest level.
Sorry for lack of details. I will post them next. Thanks.
Running Purge & Trap Tekmar ATOMX autosampler, 5mL purge volume, 11 min purge time and Supelco "K" trap VOCARB 3000. Agilent 7890B GC and 5977A MS with Agilent J&W DB-VRX 20m x 0.180mm x 1.0um column. 50:1 split ratio, Calibration range 5 to 200.
EPA 524 method accepts quadratic regression. Will this help to get acceptable RSD for vinyl chloride ?
Using the 6mm draw out plate? BFB_Autotune? Dry purge?
You should consider upping the split to 150:1.