Hello there.
It's another great day here at Noob-Chromatographers-R-Us, and I'm trying to build a method mostly from scratch. I have a published method that I'm trying to replicate, but we're a little lacking in infrastructure.
We're trying to separate some big, organic compounds, but they tend to self-destruct at higher temperatures. The good folks we borrowed an SOP from suggest using a cold on-column injector or a PTV. Unfortunately, I have neither of those available for the ol' 6890 I'd like to run this on, so I am attempting to make do with a splitless liner.
BACK INLET
Mode: Splitless
Initial temp: 100 C
Pressure: 3.6 psi
Purge flow: 0.0 mL/min
Purge time: 0.00 min
Total flow: 4.2 mL/min
Gas saver: Off
Gas type: Helium
Unfortunately, when I try to run this, the pressure is not able to get above the 1.5 psi mark. At first I thought it might be a leak, so I replaced the septum, the O-ring, and checked the liner. Tried again and it would only get to the same pressure. Opened the oven, unscrewed the column nut, changed the ferrule, and cut/reinstalled the column, and tried again. Still the same pressure.
Tried going for a splitless mode, set the split ratio at 5:1, and it gets up to the requisite pressure just fine.
What I'm wondering is, is the liner I'm using too large to pressurize?

I believe the one in there may be an SGE model 092001 that someone elected to knock the wool out of at some point. I see from the description it is "4.0mm ID Spilt, with quartz wool." Does that mean it only works with split methods? If I try wedging something inert in there to obstruct the flow, can the pressure go up higher?
Many thanks in advance for all your suggestions.
It's another great day here at Noob-Chromatographers-R-Us, and I'm trying to build a method mostly from scratch. I have a published method that I'm trying to replicate, but we're a little lacking in infrastructure.
We're trying to separate some big, organic compounds, but they tend to self-destruct at higher temperatures. The good folks we borrowed an SOP from suggest using a cold on-column injector or a PTV. Unfortunately, I have neither of those available for the ol' 6890 I'd like to run this on, so I am attempting to make do with a splitless liner.
BACK INLET
Mode: Splitless
Initial temp: 100 C
Pressure: 3.6 psi
Purge flow: 0.0 mL/min
Purge time: 0.00 min
Total flow: 4.2 mL/min
Gas saver: Off
Gas type: Helium
Unfortunately, when I try to run this, the pressure is not able to get above the 1.5 psi mark. At first I thought it might be a leak, so I replaced the septum, the O-ring, and checked the liner. Tried again and it would only get to the same pressure. Opened the oven, unscrewed the column nut, changed the ferrule, and cut/reinstalled the column, and tried again. Still the same pressure.
Tried going for a splitless mode, set the split ratio at 5:1, and it gets up to the requisite pressure just fine.
What I'm wondering is, is the liner I'm using too large to pressurize?

I believe the one in there may be an SGE model 092001 that someone elected to knock the wool out of at some point. I see from the description it is "4.0mm ID Spilt, with quartz wool." Does that mean it only works with split methods? If I try wedging something inert in there to obstruct the flow, can the pressure go up higher?
Many thanks in advance for all your suggestions.