Analysis of formaldehyde, methanol and formic acid
Posted: Thu Jun 09, 2016 7:38 pm
by OtSa
Hej,
We have got a request to analyze formaldehyde, formic acid and methanol from aqueous samples at concentrations around 10 to 500 uM. Concentration range is too low for our normal organic acid method (ion exclusion) and I was wondering if there is a way to derivatize these different functionalities either for HPLC-UV or GC-MS in water solution? If possible, I would prefer to stick with one method for these three analytes.
Re: Analysis of formaldehyde, methanol and formic acid
Posted: Fri Jun 10, 2016 1:44 pm
by itspip
I know that formaldehyde can be derivatized and analyzed by HPLC-UV, not too sure about methanol and formic acid along with formaldehyde though.
Posted: Sat Jun 11, 2016 1:53 pm
by GOM
Hi
Is example 2 in this article using GCMS of any help?
http://www.shimadzu.com/an/industry/new ... e0zdh.html
Your lower limit of 10uM for methanol would, for example, be approximately 0.3 ppm.
Regards
Ralph
Posted: Wed Jun 22, 2016 7:32 pm
by gapeev
That's a tough call to quantify them all in one run. Here is what worked well in our hands"
Sampling: SPME with Carboxen fiber at around room T
GC: Porapack-Q column starting at around 0C. Need to hold at 0 for about 2 min.
MS: Use isotopically labelled targets as IS. Monitor pertinent ions in SIM mode.
LLOQ: From low ppm to tens ppb depending on matrix: These are ions common for a lot of compounds.
Posted: Sat Jun 25, 2016 5:06 pm
by GOM
This sounds like a good approach
Regards
Ralph
Posted: Sun Jul 03, 2016 1:57 pm
by mckrause
You can derivatize the formaldehyde and formic acid but not the methanol (easily). Acetyl chloride in acetone works quite well for GC and GC-MS methods.
Methanol at 0.3 ppm is going to be tough no matter what method you use.
Posted: Mon Jul 04, 2016 9:38 am
by lmh
If you've got a good UV/vis spec, you could probably set up an enzyme-coupled assay for methanol using alcohol dehydrogenase and following appearance of NAD(P)H (340nm, roughly 6.2 mOD units per uM solution with a 1cm path-length).