Measuring of concentration by HPLC
Posted: Sat Jun 11, 2016 7:33 am
by zanio
Hi all,
I have a guestion about calculation of concentration. of amaunt of X in sample as mg/g.
It was measured on analitical balance sample with m=0,5 000g
Extraction with 15 ml solvent
Then the extract was filtrated in buchner and washed with 1-2 ml H2O.
After that 20 ml solvent was added.
5 ml of solution was taken and evaporated 1 ml.
After that 7 ml of solvend was added and put in HPLC for analysis.
The results from HPLC is 1mg/ ml.
What is the concentration in sample as mg/ g?
What is the formula that I can use?
Thank you!
Re: Measuring of concentration by HPLC
Posted: Sat Jun 11, 2016 10:52 am
by dblux_
Hi all,
Thank you!
Between 115,2 and 118,4 mg/g ( ! 1-2 mL of water !)
Was the solvent miscible with water ? No information - I assumed it was.
Never use formulas for concentration calculation. Simply start thinking logically.
Posted: Sat Jun 11, 2016 11:02 am
by zanio
Thank you!
Yes the solvent is misible in wather.
How do calculate the concentration?
Posted: Sat Jun 11, 2016 1:10 pm
by dblux_
Thank you!
1. analyte was extracted with 15 mL of solvent and 1 mL of water was added to supernatant then 20 mL of solvent was added
total volume was 15+1+20 = 36 mL
2. 5 mL was taken, evaporated to 1 mL and diluted to 8 mL and analysed
3. concentration from HPLC was 1mg/mL so in 8 mL there was 8 mg of analyte
(in fact 8 mg in 5 mL of solution - see point 2.)
4. 36 mL/ 5 = 7,2
7,2 x 8 mg = 57,6 mg in 0,5 g of sample
so in 1 g there was 115,2 mg of analyte