Back again with N-methylpyrrolidone in API
Hi all !!!..
A new year has come but troubles have not changed.
Is ther anyone who has a method to control absence of N-Methylpyrrolidone in API's. USP and EP claim taht HS/GC is not an enought sensible technic.
Any help will be appreciated.
Thanks and have a nice year !!!!!
Basil
If you can document that your procedure does meet the requirements and meets the guidelines for a validated procedure, just document it. There are updated ways to do lots of assays, and to push them through EP or USP takes time and money; lots of EP and USP procedures are long-outdated or awkward.
Hello!
We have two GC methods (it is possible also HPLC-see questions in HPLC section-we have got the answer there):
1. Column AT5 (30m x 0,53mmID x 5mcm)
FID
carrier gas: He 4ml/min
Split 1:5
Oven T:40stC (6min)-10stC/min- 150stC(5min)-3stC/min- 165stC (10min)
T inject:220stC
T det:250stC
Solvent.DMSO
We determine also other solvents so Run is long. NMP elutes after the solvent peak (RT cca 21min)
2. Column CP-SIL 5 (30m x 0,53mmID x 5mcm)
FID
carrier gas: He 5ml/min
Split 1:5
Oven T:150stC
T inject:120stC
T det:250stC
Solvent: Ether
NMP elutes after the solvent peak (RT cca 7,6min)
I hope it will help you somehow. Bye!
ana
2.
juju
If it's applicable, HPLC methodology was referenced in this thread
Thanks juju (or ana) :lol:
We will try with your first option.
I hope we would have enought sensibility with this method (if nout I'd chandge split ratio).
I'll tell you how it works in our lab
Thanks again
Basil