bupropion hcL
we use USP refer method for bupropion hcl method for RS . inwhich we use triflouroacetic acid in mobile phase a&b what is the need of TFA for mobile phase simply what role play tfa in our separation
TFA must be added as ion-pairing reagent to overcome peak tailing. Tailing factor should be less then 2 (system suitability parameters).we use USP refer method for bupropion hcl method for RS . inwhich we use triflouroacetic acid in mobile phase a&b what is the need of TFA for mobile phase simply what role play tfa in our separation
Amaryl.
@nitish: in September 2005 when I joined this forum there was a lot of discussion on use of TFA. You shall search and find the same.
@amaryl- May I get the reference stating tailing factor should be less than 2 ?
Jitender Madan
Division of Pharmaceutics
Central Drug Research Institute
Lucknow, India
Lucknow, India
Most USP methods have a limit on the tailing factor. It is commonly 2. Note that there are different ways to measure the peak tailing, and the USP prescribes a very specific one.
Here is this nice article on validation...
Shabir G A. Validation of high-performance liquid chromatography methods for pharmaceutical analysis: Understanding differences and similarities between validation requirements of the US Food and Drug Administration, the US Pharmacopeia and the International Conference on Harmonization. J Chromatogr A. 2003;987:57-66.
Check this out even ...
Dolan JW. Why do Peaks Tail?. LCGC North America. 2003 July;21(7):612-16.
Even the book by Snyder LR, Kirkland JJ, Glajch JL. Practical HPLC Method Development. 2nd ed. New York: Wiley-Interscience; 1997.
Amaryl.
