siRNA separation
3 posts Page 1 of 1
the duplex of siRNA
my iclass
column temperature 25
TEAA 100mM
mobile phase A
mobile phase B 20% A in ACN
gradient elution 35% --85%
anyone got the experience of oligonucleotide analysis ?
Typically one analyzes single-stranded oligonucleotides smaller than ~ 18 bases by reversed phase and by anion-exchange if they're larger. Anion-exchange is the method to use for double-stranded oligos. What exactly are you trying to separate from what?
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
3 posts Page 1 of 1
In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 11 guests