Hydrogen detection
Hi
Why am I not able to detect Hydrogen using HaySepC and molecular sieve 5A columns with TCD detector, Helium as carrier gas,
Injection temperature = 200C
TCD Temp = 200C
thanks
Suni
You did not mention the concentration at which you are trying to detect and the amount of gas you are injecting.
There are a lot of reasons why you are not detecting H2.
Detector problems
Leak problems
Injection problems
The low and NEGATIVE response H2 has using He as a carrier on a TCD detector
I hope this will give you enough information to troubleshoot your analysis problem.
Good luck,
Thanks, Rod.
I have checked the leak,etc..
I tried with 5% Hydrogen in Nitrogen.
The sample flow rate in GC is 15ml/min.
Also using 10 port gas sampling valve.
thanks for the immediate reply....
regards
Suni
H2
You did not state what column you are using. A poor quality column or a poor installation will make H2 difficult to see at only 5% using Helium carrier.
Try using Nitrogen carrier for improved sensitivity.
good luck,
Two columns - HaySepC and molecular sieve 5A
thanks
Suni