Agilent HPLC manual injector
I have an Agilent 1100 HPLC and I've just bought a Rheodyne 8125 manual injector.
My problem is that I can't get it to send the start signal to the Chemstation software. I'm new to Chemstation, and I've been looking for a setting that tells it to look for a manual injection, but can't find anything.
I don't have the factory cable - instead I've made a cable up that hooks into the APG port with a BNC connection. The position sensor has been checked and is open circuit in "load" and closed in "inject", right up to the connector.
If someone can help, I would greatly appreciate it.
Is this a software setting problem, or a hardware problem?
Hello
Please follow instruction below:
1.Go to “method and run control” mode in Chemstation
2.Go to: “Instrument” – “More injectors” and click “use external injector” (you should see mark symbol next to it "✓")
When you start your run Chemstation status will be “pink – waiting for injection”. Of course you need remote cable to connect external injector device with LC module. Otherwise you won’t be able to trigger your LC automatically.
Regards
Tomasz Kubowicz
Users browsing this forum: Semrush [Bot] and 14 guests