Waters 996 Photodiode Array Detektor
2 posts Page 1 of 1
a completely "hplc beginner" needs help :-)
how can I check the lamp intensity of a Waters 996 Photodiode Array Detektor?
Login the Empower Pro interface. Find the Diagnostic/PDA Calibration menu in the Run Sample window. You can see the spectra of the lamp. Flush the system with degassed MeOH and set up the exposure time of the diodes 21 ms. The max. of the spectra about 230 nm is 30000-40000 with new lamp and good optics.
2 posts Page 1 of 1
In total there are 10 users online :: 1 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot] and 9 guests