Reverse Phase Buffer B
Posted: Wed Jan 04, 2017 2:25 pm
Maybe I'm daft or maybe I missed a detail somewhere. Most of knowledge comes from methods papers, reviews, and book chapters in my field of interest and if that matters it's nucleic acids. I only mention this because I've never read a pure HPLC theory book.
Anyway, buffer B is usually listed as ratio of A(aq):B(org). Occasionally the author will specifically detail that X L of A was mixed with Y L of B. So if you're using an A buffer of 10 mM ammonium acetate (AA) and you make your 50:50 buffer B by mixing equal volumes of A with acetonitrile, your buffer B's final composition 5 mM AA and 50% acetonitrile. A gradient from A to B is a gradient of increasing organic but also decreasing buffer. Is this intended?
My buffer preparation entails adding desired amount of AA stock to a volumetric flask and then proceed. For A, I fill to 90% with water, invert a few times, sonicate, then fill to to 100%. For B, I add enough water for the desired final % aqueous of B, slowly add my organic in portions whilst inverting to 90%, sonicate, and finish to 100%. Such that, both my A and B have same concentration of buffer and is not variable over the gradient.
Is this incorrect?
Thank you.
Is this incorrect?