ionization stability on API 2000
Dear friends
I am working with determination of mycophenolic acid in plasma samples by LC-MSMS.
I developed a method to exctract and optimized the MS.
I have a variation of ±10% between injections (n=7-15) of the same vial no matter if I test pure standard or extracted sample.
mobile phase> acetonitrile-acetic acid 1mM (8_2) isopropanol 5%
ion source temperature> 450°C
flow 0.8ml_min
split 1_1
Does anybody has an idea about the reason of this variation ? I am not working with my limit of quantitation value in other words, it is not a matter of low concentration.
Hope to hear from you soon......
Hi Maristela
When you say you have a variation of 10% do you mean the absolute peak areas or the ratio to the internal standard. What internal standard are you using? mycophenolate can be difficult if you use the wrong one.
Geoff
[size=75][color=blue]There's a saying amongst prospectors 'Go out looking for one thing and that's all you'll ever find'[/color][/size]
In total there are 31 users online :: 1 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Users browsing this forum: Google [Bot] and 30 guests
