Obstruction Superdex 200 in FPLC system
Good morning,
I have been using a Superdex 200 in the last step of protein purification. I use a very old fplc system (Pharmacia Biotech). After 8 injections of 1 mililiter (10mg protein) the pressure was high even at low flow rate. I tryed to clean the column inverting it and following the cleaning in place protocol from GE Healthcare user manual: 25ml 0,5M sodium hydroxide at a flow rate of 0,5ml/min. I charged an old 10ml superloop to inject. As the pressure was high I had to lower the flow rate to 0,1ml/min. I suspect that the hydroxide degraded part of the loop and all that entered the column as I observe the pressure does not lower. I washed with 3 column volumes of MiliQ water. Pressure its still very high, reaching the max permited at a flow rate of 0,1ml/min.
Please, what do you recommend doing? We cannot afford buying a new one.
Thank you very much in advance for your help.
Luciana from Argentina
Replace the column.
Before replacing the column, just disconnect the column and notice the pressure. If it is still high (0,2-0,3 MPa, high for FPLC) it might be a inline filter of the pump that is clogged and needs to be replaced....a lot cheaper then a new Supderdex Column.